I've got an issue that appears to be caused by damaged resistors that are either open circuited or of too low a value due to contamination. The problem is that they're gigaohm resistors, so to a multimeter, they're always open-circuited. How can I measure the resistance, or, at least, test the continuity?
|
|
Many Fluke meters (e.g. 87,287) have a nanoSiemens conductivity range which will measure up to 100 GigaOhms - it needs to be manually ranged up from the ohms range. \$\mathrm{1 G \Omega = 1 nS}\$, \$\mathrm{10 G \Omega = 0.1 nS}\$. Alternatively, most DMMs have with a 10M input impedance (easily checked with a second meter), so a resistor with value R in series with the millivolt range will form an R+10M/10M voltage divider. So applying 10 volts through a 1 gigohm resistor will read around 99 millivolts. A close-enough approximation for high value resistors from a 10V supply would be resistance in gigohms = 100/millivolts. |
||||
|
|
|
You need insulation testers. The ones I've seen had 2 GOhm range. Not necessary Flukes, there are cheaper ones. And for the future, I would try to add some protective insulation on top of such nasty things :-) |
|||
|
|
|
I'll assume that you're able to isolate the resistor from the rest of the circuit. You probably need to construct a high-impedance analog buffer. It doesn't need to be super-fast, but it does need to be high impedance. A very high impedance amplifier is National's LMP7721, requiring only 3 femtoamps of bias current. Once you have your buffer, get another resistor with a resistance comparable to the one you want to test (a known value). Connect one side of this resistor to ground, and the other to a probe and to your buffer. Then, apply a voltage to one side of your resistor, and connect your buffered probe to the other side. Measure the voltage at the output of your buffer and solve the voltage divider to determine the unknown resistance You may not need a buffer if your meter has extremely low impedance when measuring voltage. |
|||||||||||
|
|
There is special equipment for this. Couple of weeks ago someone showed me one that can do > 500G and in this particular case was used to test 10kV breakers. It was called a Megger. Basically what it does is measure resistance, but where your multimeter dus this with 3V, these thing slowly increase the voltage to test in the range of kV's. https://en.wikipedia.org/wiki/Megger I expect there are other vendors for similar equipment. |
|||
|
|
|
"If you use a battery operated DMM, and keep it isolated, you can use 1000s of volts for the test." DONT TRY THIS !!! Most of GigaOhm resistors, including 200 GigaOhm resistors in glass tubes have a rating of maximum 500 volts, and the maximum voltage for a digital voltmeter is 1000 volts. Thousands of volts across such a resistor will only be sparking around the resistor and instantly fry your digital voltmeter! |
|||||
|
|
What you would want is a megaohmmeter. These are just another permutation of the If you use a battery operated DMM, and keep it isolated, you can use 1000s of volts for the test. I used to calibrate the leakage current readings of 1-200KV Hi-Pots using just a normal fluke DMM with this method. You can find megaohmmeters on ebay as "Hi-Pots", "insulation tester", "oil tester", "dielectric tester". Also, the opposite of a megaohmmeter is a digital low resistance ohm-meter(DLRO), these use a high current (1-100+ amps) to measure very low resistances. |
||||
|
|
|
I have just tried to measure 10 Gigaohm resistors with my DMM and a 10 volt power supply with success. My DMM is a 4 1/2 Digit with a stated 10 Megaohm imput impedance. The DMM has an accurancy of 0.05 % for voltages measures. I first adjusted my power supply so that the voltage shown an my DMM was exactly 10.000 volts, then put the 10 Gigaohm resistor in serial with the DMM on it's 200 mV range. The reading was 11.35 mV. In fact, the only thing that is not stated as precise with my DMM is it's imput impedance! I tried to measure it with another multimeter (not digital) and found that the real imput impedance of my DMM is in fact above 11 megaohm, so there are about 10 % error. The 10 Gigaohm resistors I measured (I have 4 of them) have only a 5 % tolerance, but they all gave me about the same reading on my DMM. If I had one of 0.1 % tolerance, I could adjust my power supply so that the DMM would read exactly 10 mV to compensate for it's 11.35 Megaohm impedance, in this case the voltage from the power supply would be adjusted to 8.81 V and I would have a precise gigaohm meter. Another thing to note is that the probes of the DMM have lot of leakages. I had to put the DMM on a separate table with the probes and the resistor to be measured hanging on the air. I then tried to put the 10 volts from the power supply across the PVC part of each probes and had a voltage reading of 0.05 mV on the DMM, corresponding to a resistor of about 2 Teraohm... Time to buy teflon insulated wires... |
|||
|
|