A piece of measurement equipment which can visualize one or more signals simultaneously on a time scale.

learn more… | top users | synonyms (1)

1
vote
3answers
176 views

How do I average a complex waveform?

I am trying to figure out how long my project will last on a single battery. It's powered by a single AA battery with an NCP1402 3.3 volt boost converter. I've determined that the target circuit draws ...
0
votes
0answers
25 views

How to measure the peak current (inrush current) and results assessment

I would like to peak the inrush current of a controller. The voltage input of the system is 24V and the only thing which is available, is an oscilloscope and a variety of resistors. The basic idea is ...
0
votes
2answers
75 views

Difference between DMM and oscilloscope measurements

I have a 220 - 14 V transformer, but when I hook it up to oscilloscope or DMM they show different values. DMM shows 11.8V and oscilloscope shows 17.2. Ok, this is cheap Chinese transformer, so it can ...
0
votes
1answer
99 views

Oscilloscope ground and unit under test ground issue

I am working with an instrument which has a supply of 230V AC or 400V DC. AC source is isolated at the input via 1:1 transformer and DC source is a floating supply. I have a Yokogowa DL1620 ...
0
votes
0answers
82 views

What causes this FFT artifact in my DSO?

I am curious what causes this (apparent) transient anomaly in my DSO's FFT display. The pattern is easiest to see if I ground the probe tip to the ground clip. Then the time display is only a solid ...
1
vote
3answers
138 views

Getting the right frequency from an oscillator circuit

I have constructed an oscillator circuit using the resistor and capacitor values shown in the diagram above (from the Proteus ISIS simulator). I run the circuit and use a switch to trigger the ...
2
votes
1answer
38 views

Gas Discharge V-Q Lissajous

I've been investigating non-thermal plasmas, specifically dielectric barrier discharges, where two electrodes, at a potential difference of several kilovolts, are separated by a dielectric and a gas ...
0
votes
1answer
66 views

Nature of noise gotten from digital oscilloscope

I've got strange noise while using digital oscilloscope (Auris B-424). Spectrogram of this signal presented on this image: (X time in seconds, Y frequency in Hertz, in this image case from 0 to 50 ...
2
votes
2answers
93 views

Measure 230V mains with 300V CAT II probe?

I need to measure harmonic distortions on a 230V 50Hz AC mains circuit. The Tektronix P2220 probes I have for my scope are rated for 300V CAT II when set to 10X impedance. Are these probes safe to use ...
2
votes
2answers
186 views

Can an oscilloscope add a frequency component to my signal?

I want to check the spectrum of a DDS signal ("pure" sinewave) for undesired frequency components. The way I am doing it is by storing the waveform from a 2GS/s oscilloscope (Agilent DSO81204) and ...
-5
votes
2answers
51 views

the circuit of active probe using FET trace CRO [closed]

pls suggest the circuit of active probe using FET trace CRO.
2
votes
2answers
74 views

Continuously log data from an oscilloscope

I have a pin which is being toggled every 100ms by an MCU. I need to plot a graph which consists of the variation from the desired interval . That is the number of times it moved away and was toggled ...
0
votes
1answer
104 views

Capturing PWM Signal on an Oscilliscope

I have a stepper motor that is controlled by PWM. I am trying to analyze the signal coming out of the controller on an oscilloscope. Ideally, I would like to capture, maybe 2 seconds of wave form so ...
1
vote
3answers
351 views

Why does an oscilloscope show a different voltage from a signal generator?`

I have a Rigol DS1052E oscilloscope and a Rigol DG1022 function generator. I have just changed the oscilloscope fan for a silent one, as I found, as is commonly reported, that the original fan noise ...
2
votes
1answer
55 views

How to store more waveform after trigger and less before?

I have a Rigol DS1074Z and can't figure out how to lower the time the scope spends pre-sampling to open up more memory for the actual data I want to see. Here is a nice image that will hopefully ...
0
votes
2answers
83 views

Eye diagram test

I am trying to perform a eye diagram test on a device to check its USB performanance. I am using a board "SQIDD" in order to perform this test. The following figure shows my test setup, I am using ...
4
votes
2answers
472 views

Why does reversing the oscilloscope probes cause ringing in the signal?

I am measuring a PWM signal in an oscilloscope using probes with a 1X and 10X switch (active?). I'm measuring across a resistor which is in series with an LED which is driven by an NPN 2N3904 ...
0
votes
1answer
88 views

How to calculate RMS of total noise from digital Oscilloscope (FFT)?

Only the red channel is in consideration. The yellow curve is not relevant, which is only another channel of the oscilloscope. The image above was captured from a digital oscilloscope. The vertical ...
1
vote
3answers
156 views

Designing a test instrument to be connected to an oscilloscope

I'm working on a piece of test equipment that is intended to be hooked up to an oscilloscope to measure the waveforms it produces. Since that's its reason for being, I figured it made sense to provide ...
4
votes
3answers
1k views

What will I see in oscilloscope when attempting to analyze Bluetooth or Wi-Fi signals?

I would like to know what will show up if I put bluetooth or wifi carrier signal (2.4GHz family) in a say 50MHz oscilloscope? Am I going to see something like a rectangular block? And the only ...
1
vote
1answer
80 views

See if signal is amplified without an oscilloscope

I'm an electronics newbie, but I've had some experience with oscilloscopes in college lab. However, I need to design and wire a simple circuit that can read values from an electret mic. Since the ...
1
vote
5answers
223 views

Can I see an analog signal with changeable frequency on oscilloscope?

I have a sinusoidal signal with changeable frequency (which implies that it is not periodic). I wonder is it possible to see it on oscilloscope? Does it show non-periodic signals? As it is apparent my ...
2
votes
2answers
293 views

A way to measure unknown high voltages with an oscilloscope

I have a sort-of black box of a circuit that is used to generate plasmas. I'm sure it outputs on the order of kV but I'm not sure at all about the current. I was wondering if there is a way to measure ...
2
votes
2answers
91 views

Obtaining O-Scope signals in a AC/DC supply - grounding issues

I need to check waveforms in the power supply from this design reference by Texas Instruments. I essentially built the supply, but it is not operating correctly - ie, no output current, but it has ...
0
votes
1answer
107 views

Put a DAC at the output of FPGA

I have designed a circuit by System Generator to implement on FPGA. The output signal is a sinusoidal with changeable frequency. I need to read the output signal by oscilloscope. I should put a DAC at ...
3
votes
3answers
99 views

Oscilloscope gear and ESD

The answer might be obvious, but I am curious, how well are oscilloscopes and probes usually protected against ESD? I noticed that oscilloscope probes are usually shipped in esd-protected bags. On ...
0
votes
1answer
76 views

Make sure this equipment can do what I want (infrared decoding) [closed]

I'm CS but took a digital electronics physics class at my old university and found it really interesting so I wouldn't mind having my own small EE workshop. For my main project I just want to decode ...
2
votes
1answer
71 views

oscilloscope time charging different from the math

For the above circuit, I calculate that the equivalent resistance of the charging circuit is 2.668 kΩ, and the equivalent capacitance is 2.055 µF. Therefore, the time constant would be ...
0
votes
1answer
138 views

TV signal into an Oscilloscope

I'm plugging TV signal cable directly into an oscilloscope. TV cable looks like this: First I connected outer shell of TV cable connector to GND of oscilloscope BNC cable and center to the center of ...
0
votes
1answer
116 views

BNC cable instead of oscilloscope probe for fixed setup?

I'm developing a device, it will generate a signal that will be fed directly to an oscilloscope, and connected permanently. I now have the circuitry set up on a breadboard, attached to the scope ...
0
votes
1answer
77 views

Oscilloscope pulse frequency

I'm measuring pulses from a device with a constant frequency. The oscilloscope's usual frequency indicator shows 600 Hz. But when I use "Measure" button and then Type is set to Frequency, then that ...
1
vote
1answer
539 views

why is this oscilloscope so cheap with a 20Gsa/s sampling resolution?

I've looked around this site but I haven't been able to figure out why this oscilloscope seems to have such a high sampling rate "Repetitive mode: 20GSa/S" and be relatively cheap: ...
5
votes
2answers
1k views

How to test a crystal resonator in board?

I have 2 quartz crystal resonators in my pcb board: 32.768 kHz & 20 MHz. They are connected to a Freescale MC12311 transceiver IC, which has an HCS08 micro-controller embedded in it. I want to ...
1
vote
2answers
68 views

Is the voltage from peak to peak my thevenin in voltage?

I used a function generator and set it to 1kHz. And used an oscilloscope to measure the time constant of an rl and rc circuit. my question was that what would I consider as Vth or voltage thevenin , ...
0
votes
1answer
90 views

What are the demo 1 and demo 2 terminals on the oscilloscopes?

I have seen almost all oscilloscope providing this demo1 or may be demo2 terminals on their oscilloscopes which has the square wave running all the time? I can't figure out the purpose of those demo ...
3
votes
2answers
181 views

Oscilloscope sweeps only once with DC sources and shows nothing on GND setting

I got an old CRT oscilloscope someone had laying around unused in his garage (an ATAIO AI 751A which seems pretty obscure, not getting any meaningful Google results). Unfortunately the user manual is ...
3
votes
2answers
154 views

Measure peak voltage / amp, when signal is very short (relay on/off)?

I was measuring a 'steer' signal for switching on a relay. The oscilloscope, gave me back it peaks at around 2.64V. The width of the signal is 132ms. I then hooked up my multimeter, to measure how ...
4
votes
1answer
107 views

How can I track where this 471kHz noise is coming from?

I'm new to EE in general and I just received my first oscilloscope (Rigol DS1102E). Anyway, I'm building an audio circuit using an Arduino Micro and the infamous ...
4
votes
1answer
148 views

Why does my crystal oscillator spike so much?

I just got a new oscilloscope (Rigol DS1102E) and I'm new to EE in general. One of the first things I put on the scope was my 3.579Mhz "TKD" crystal oscillator. ...
1
vote
2answers
147 views

Filter 50Hz hum picked up by the oscilloscope

I am measuring huge 50Hz oscillations on my oscilloscope, but I am not entirely sure the probe is generating those by behaving like an antenna (an ADC at the same point gives stable values but it's ...
1
vote
0answers
50 views

Acoustic signal on oscilloscope [closed]

I'm trying to measure some echo acoustic signals in an oscilloscope DSO-X 2002A 70MHz bw, 2GSa/s from discontinuities in a rock cylinder. The acoustic velocity for P-waves (which is the type of ...
2
votes
1answer
118 views

Effect of the high impedance or low impedance of oscillospe

I am measuring a sine waveform range 0 to 40MHz by using Agilent InfiniVision DS0-X 3034A. The data sheet of the oscilloscope is here: http://cp.literature.agilent.com/litweb/pdf/5990-6619EN.pdf My ...
0
votes
1answer
141 views

AC and DC voltage are constant even though the frequency changes

In the circuit below, why does the AC voltage remain constant even though the frequency increases?
5
votes
2answers
872 views

Terminating a 50 Ohm source into 1M Ohm oscilloscope input

Due to buying some equipment several months apart I am now faced with the task of connecting a high speed ,up to 1GHz, light detector that requires a 50\$\Omega\$ load impedance, but my oscilloscope ...
0
votes
2answers
71 views

Choosing Oscilloscope with appropriate bandwidth for Jitter analysis

Can someone give me a rule of thumb of choosing Oscilloscope with appropriate bandwidth for the Jitter analysis of clock signals? What device is highly preferable: spectrum analyzers, frequency ...
3
votes
2answers
321 views

Cannot get a digital square wave on oscilloscope from the GPIO of the Raspberry Pi?

ENVIRONMENT: Running "Raspbian" on Model_B Raspberry Pi (Revision 2) w/ ribbon expanding the GPIO out onto a Breadboard as shown in the following image. SETUP Here is a close up of the GPIO ...
4
votes
1answer
127 views

Can the DS1052E be used for reliable NTSC measurement?

I'm in the market for an oscilloscope. I've read that the infamous DS1052E from Rigol is a great beginner scope due to the price/performance. However, I know I have a lot to learn and simply don't ...
1
vote
1answer
57 views

View High Voltage AC Signal on Oscilloscope

I would like to capture a high voltage AC signal on an oscilloscope. Input signal: 160V @ 400Hz Output signal: ~40V @ 400Hz Oscilloscope: TDS3014C I intend to use a voltage divider: 100k ohm (1/2 ...
1
vote
1answer
62 views

Measure both sides of an AC and DC circuit

So I have recently begun messing with diodes and there use in converting AC to DC. What I am curious in trying to do is to hook up my dual channel oscilloscope on bother DC and AC side to view the ...
2
votes
1answer
100 views

Ground noise measured on scope

I've connected a scope probe directly to the ground clip of the scope. I then observe the signal on the scope and it's reading 0, as expected. I set my trigger to be 1V above the 0V reading it ...