I have a circuit (shown in figure 1) that is supposed to function as a precision analog output. It outputs a DC voltage in the range of either ±10V or ±5V, depending on which gain resistors are used. This analog voltage is intended for use as a reference type of voltage, entailing:
- High Accuracy ±(0.1% + 1mV): after calibration
- ~5-10mA of output current maximum
- DC output, no transient slewing specifications formally stated
Figure 1: circuit in question.
For +/-5V output the gain is \$G=1+\dfrac{R_1}{R_2}=2\$
For +/-10V output the gain is \$G=1+\dfrac{R_3}{R_4}=4.026\$
Output is \$G\cdot(V_{DAC}-2.5V)\$
After a calibration, the output of the circuit performs to spec. The calibration I used is a simple slope and offset error correcting equation that operates on the code I send to a DAC, shown the code below.
voltage = 5.0; // set voltage
cal = (voltage * 1.0047) + 0.0144; // slope and offset calibration
value = ((cal + 5) * (65535 / 10)); // translate -5 – 5 –> 0–5
if (value > 65535) // send max output if value > 2^16
value = 65535;
output_MAX5134(value, AO_DAC_0); // write to DAC output 0
The problem is, if I come back the next day after calibration, there will be an offset present on the output. The gain error is still eliminated, but I will see something like 2.5200V instead of 2.5005V. The thing I can think of is the temperature is effecting things, maybe I have to account for it?
For reference I am using:
1% Tolerance SMD resistors -> should be accounted for in calibration
REF3425 -> 0.05% 2.5V Reference
http://www.ti.com/lit/ds/symlink/ref3425.pdf
INA145 -> Difference amplifier as Unipolar to bipolar converter
http://www.ti.com/lit/ds/symlink/ina145.pdf
MAX5134 -> 16-bit DAC with high linearity
https://datasheets.maximintegrated.com/en/ds/MAX5134-MAX5137.pdf
I am using a u3606A 5-1/2 digit multimeter for voltage measurements
INA145 Gain Error Vs Temp:
INA145 Offset Voltage Drift Vs Temp: