We're using a TI LMP90080 as an ADC to read a 4-wire resistance thermal detector (RTD) in the following circuit: 4-wire RTD
The problem we are having is that when the ambient temperature changes, especially near R88, a 25ppm/°C 0.10% 1k resistor, the reading is negatively effected, i.e. when heated, the temperature read changes about 20 °C negatively, and vice versa when cooled. This seems to make sense but I'm wondering is there a way to fix this without temperature compensation via an ambient temperature IC?
The RTD is connected via traces and some interconnect wires between PCBs and are terminated as such:
termination
The RTD is well away from the heating or cooling used for testing and looks like this:
rtd-wire
  • \$\begingroup\$ What's a "RTD"? \$\endgroup\$ Jun 12, 2014 at 16:14
  • \$\begingroup\$ Good point, I'll fix that. \$\endgroup\$ Jun 12, 2014 at 16:15
  • \$\begingroup\$ How much is the RTD reading affected by the changes in ambient temperature? How is the RTD wired (what type of wire, how long, a photo wouldn't hurt)? What is the temperature coefficient of R88? \$\endgroup\$ Jun 12, 2014 at 16:27
  • \$\begingroup\$ Of course RTDs are sensitive to temperature, that's what they are designed to do. (more serious answer below). \$\endgroup\$ Jun 12, 2014 at 17:15
How the heck do you figure this "makes sense"? 25ppm/°C is for a 20°C change is about 500ppm, or about 0.05% (and that's the maximum change, typical is usually better.
A standard Pt RTD changes about +0.385%/K so you should not get much more than about 0.1°C change in your reading.
What could be happening if you're using some crude means like a hot air gun to heat the board is that you could be seeing thermocouple voltages, but even that seems unlikely, unless you're doing something particularly silly.
Since it appears that a hot air gun was involved, let's explore that angle. Say a temperature gradient of 60°C is created by the gun (a cheap one will produce air at hundreds of degrees C). If a junction between two dissimilar metals are present will produce 20 or 30uV/K then you could get 1.5mV of voltage. A DIN-standard Pt100 RTD at 1mA will have about 0.38mv/K output, so that could cause 4 degrees C error. The reported 20°C is possible, but they'd have to be total blacksmiths (no offense intended to actual skilled blacksmiths). You can't really get realistic results with that sort of test- a proper thermal chamber with controlled slew rates should be used (and, as I said, you should see only very small shifts). Presumably something is wrong that is causing the client to do these things. If I was a betting person, I'd guess bad wiring, electrical noise (your circuit is not industrially hardened, but it could be okay in a lab), or possibly a mismatched sensor
Use a voltmeter to measure the input voltages wrt ground and make sure they are reasonable, then do the calculation manually to figure out what is going on. Make sure the current through the RTD is high enough that thermocouple voltages won't be a huge effect (don't set the current to 100uA with a Pt100 RTD). It's a trade-off between self-heating and having enough signal to work with.
Something is seriously wrong to be seeing changes like that with such high-end parts-- it needs to be fixed, not compensated.
  • \$\begingroup\$ Regarding it making sense, I meant the direction of temperature change, not so much the values. Anyways, I didn't consider the amount of current through the RTD. According to this, 1 mA is recommended and that is what I have set within the LMP90080 (it doesn't go higher than that but it can be much lower). I agree that it should be fixed if something is wrong though. \$\endgroup\$ Jun 12, 2014 at 17:53
  • \$\begingroup\$ Temperature coefficient of a precision resistor can go in either direction, change, etc. It may be 'U' shaped. 1mA should be fine, please check voltages. \$\endgroup\$ Jun 12, 2014 at 18:06
  • \$\begingroup\$ I tried learning some about thermocouple voltages but couldn't find anything that made sense with how you were discussing it. The issues was first noticed when our client put a hot air gun to the board and recorded the ambient temperature going up and the reported RTD temperature going down. The voltage at RTD1 and RTD2 were about 1.92V and at RTD3 and RTD4 where about 1.82V. \$\endgroup\$ Jun 12, 2014 at 20:15
  • \$\begingroup\$ I'll expand my answer a bit to cover the thermocouple voltage angle. \$\endgroup\$ Jun 12, 2014 at 20:41