In a data acquisition system absolute accuracy is defined as measure of all error sources as following:

Abs. accuracy = error from Gain(Span error) + error from Offset + (error from Noise + Quantization)

Here in the following data sheet at page 30:

http://www.mccdaq.com/pdfs/manuals/PCI-DAS6034-35-36.pdf

for analog inputs the absoulte accuracy for +-10V range is defined as 10.2*LSB where I calculate absoulte accuracy as:

10.2 * (20/2^16 = 0.00030517578125V) = 3.11mV

And at page 31 there are absolute accuracy components at table 2:

Offset err: 531µV

Noise + Quantization err: 180µV

Absolute accuracy at Fs: 3.1mV (This must be the gain error)

When I add up these three it makes 3.81mV instead of 3.11 expected.

What am I doing wrong here?

\$\endgroup\$
4
  • \$\begingroup\$ My guess would be that absolute accuracy is absolute accuracy. 3.11 = 3.1 within rounding error \$\endgroup\$
    – PkP
    Oct 13, 2015 at 8:29
  • \$\begingroup\$ Oh ok so the total error should be 3.81mV right? \$\endgroup\$
    – user16307
    Oct 13, 2015 at 9:12
  • \$\begingroup\$ Am bit confused, so abs. accuracy do not include offset and quantization erros? \$\endgroup\$
    – user16307
    Oct 13, 2015 at 9:13
  • \$\begingroup\$ I think sweber got it right in his answer; I also think that abs. accuracy is the sum of all errors (gain + offset + noise). \$\endgroup\$
    – PkP
    Oct 13, 2015 at 9:46

1 Answer 1

5
\$\begingroup\$

You can not measure the accuracy in ADC steps directly. Instead, you measure linearity, offset and noise error in volts, because your reference system gives you volts, and the device under test (your DAQ) presents its measurements in volts, too.

The result of this measurement can be found in table 2 on page 31 of the document linked by you.

It states, that at +/-10V range, the error is:

0.0239% of current reading (gain error). At 10V, this is 2.39mV
0.531mV offset
0.180mV noise

The sum of this is 3.101mV, which is 3.10V when rounded (right column in table).

To interpret this values, they are compared to the theoretical resolution, which is limited by the ADC resolution.

Your ADC has a resolution of \$\frac{1}{2^{16}-1}\cdot 20V\$. Dividing the values above by this resolution gives

10.16120175 LSB for 3.101mV (rounded: 10.2 LSB)

and still

10.157925 LSB for the rounded value of 3.10mV (rounded: 10.2 LSB)


One side note: Noise is a statistical error, which means that each measurement differs from the true value. Often just a little, but sometimes also more. This distribution usually follows a gaussian function, and the relevant parameter is the standard deviation. So, if your noise error, which is 0.6 LSB, is a standard deviation, this means 66% of all measurements of the same voltage will be contained in a range of +/-0.59LSB, so in a range of 1.18 LSB width. 90% of all measurements will be contained in a range of twice this size.

So, don't pay too much attention to the last decimal of the absolute accuracy.


EDIT:

Due to your comment, here's a bit more on how to interpret the different values for accuracy. Imagine a quite lousy multimeter, to which you apply precisely known reference voltages from 0 to 10V, and you read the multimeter 20 times for each of this voltages. If you generate a diagram of this data, you get something like this:

enter image description here

For a perfect multimeter, all points should reside on the green line, because

$$U_{measured}=U_{reference}$$

But here, this isn't the case: The average voltage measured by the device reside on the blue line. There is an offset of +2, and the slope is 1.5, i.e.

$$U_{measured}=1.5\cdot U_{reference}+2$$

The noise in this case follows a gaussian distribution with a standard deviation of 1. This means, about 66% of all measured voltages are within the area between the bold orange lines, and 90% are between the thin orange lines. Or:

66% of all measured voltages are in a range defined by

$$U_{measured}=1.5\cdot U_{reference}+2 -1$$

and

$$U_{measured}=1.5\cdot U_{reference}+2 +1$$

The last formula also is the worst-case scenario: Slope and offset go into the same direction (upwards), and we add a +1 for the noise.

Finally, offset and noise are constant, only slope error depends on the measured value.

Note that I assumed a gaussian distribution of the noise and the standard deviation as parameter. It is possible that the distribution is different and that a different parameter is stated in the data sheet. But what I said is still true: The noise defines a boundary in which most / all measurements reside.

\$\endgroup\$
6
  • \$\begingroup\$ great answer but just one point, you wrote "The sum of this is 3.091mV" isnt the sum 3.101mV? \$\endgroup\$
    – user16307
    Oct 13, 2015 at 9:58
  • \$\begingroup\$ @user16307: Indeed. I'll edit it soon. This even reduces the rounding error. \$\endgroup\$
    – sweber
    Oct 13, 2015 at 10:01
  • \$\begingroup\$ one more question. you mentioned the noise is statistical. for gain and offset error the table values are for Full Scale. so if i want to calculate the error for 5V DC input, would it correspond to abs error = 0.0239%*5V + 0.531mV + noise error ? so only the gain error is proportional to the input voltage right? so the true value will be = measured value +-abs. error. but would you document it as uncertainty or subtract it for accuracy? \$\endgroup\$
    – user16307
    Oct 13, 2015 at 11:54
  • \$\begingroup\$ @user16307: I've expanded my answer, I hope it's more clear now. \$\endgroup\$
    – sweber
    Oct 13, 2015 at 12:58
  • \$\begingroup\$ yes more clear now, but the table says all errors are +-. if you would lets say measure with this daq board 5V, would you state that the true value is (measured value +-abs. accuracy) or would you subtract the gain error and offset error and only document the +-noise error. i mean are those gain and offset errors fixed or are they max values? im asking because if the gain error is proportional to applied voltage and offset, why daq doesnt take care of it before sending to serial port?? \$\endgroup\$
    – user16307
    Oct 13, 2015 at 14:40