I have an instrumentation amplifier which output voltage equation is given.The piezoresistive pressure sensor in Figure E3.12 consists of a bridge of four silicon strain gages of 4000 ohm and 1 mV maximal offset when supplied at 12 V. If the op amps are supplied at 0 V and 5 V, I need a pressure range from 0 psi to 100 psi to yield an output voltage from 0.5 V to 4 V. My question is, why should I add the factor Vb/12 when calculating for 4V?enter image description here
  • \$\begingroup\$ The factor is determined by the values of several resistors in the circuit. Unfortunately, you haven't included those resistor values in the schematic. \$\endgroup\$ – Dwayne Reid Jul 3 '16 at 16:28
  • \$\begingroup\$ The resistors are not given. I must design them. In this case Vb/12V in the equation for the 4V is given before the resistors. I need to know why that Vb/12V is there. \$\endgroup\$ – user115094 Jul 3 '16 at 16:34
The strain gauges in this circuit are not used in a constant current arrangement, but are supplied by a constant voltage source (Vb) instead. I think the 1mV/psi response you mention has been determined for a supply voltage of 12V (which you also mention in your question).
Since the bridge is operated in a constant voltage configuration, if you supply the bridge with something other than 12V, it's voltage response will be proportional to the supply voltage by the factor Vb/12.
Said differently: the current in each leg of the bridge will be nominally Ih = Vb/(2*R), where R is the nominal resistance of each bridge resistor. If the strain changes only one bridge resistor by ∆R, then the voltage response of the bridge would be v1-v2 = Vb/(2*R)*(R+∆R) - Vb/(2*R)*(R) = Vb/(2*R)*∆R, so it's obvious from this equation the bridge voltage response is proportional to Vb.
Incidentally, the bridge offset (1mV @ 12V) will also be scaled by Vb/12.