I'm trying to diagnose a faulty current limiting circuit in a piece of lab equipment from back in the 80s. The fault is too much current.
The tail end of the circuit is below. Q3 represents a complex control/feedback circuit, and R3 is the test load. For the purposes of testing I've tied Q3 base low. Thus the emitter-base voltage of Q1 is 0V.
My understanding is this should shut down the darlington, but I'm seeing a base-emitter voltage on the darlington of approximately 2v.
Does this suggest Q1 or Q2 is faulty? Is there any way of testing the darlington without pulling it out of circuit?
    \$\begingroup\$ If you pull the base of Q1 to ground, then Q1 will conduct. 2Volts then sounds pretty reasonable for the base of Q2. That's 2* 0.7 Volts for the two base/emitter junctions in Q2 plus the voltage drop across R3. \$\endgroup\$ – JRE Jan 23 '17 at 6:27
    \$\begingroup\$ I've pulled the base of Q3 to ground, so this should push the base of Q1 high? \$\endgroup\$ – scrooge Jan 23 '17 at 6:28
  • \$\begingroup\$ @JRE: not the base voltage of Q1 is 0V but the emitter-base voltage. \$\endgroup\$ – Curd Jan 24 '17 at 10:06