I have been reading about various methods to perform a Cable Path Loss for RF Testing but not being a core RF person, I needed some insights on how it is done in Field or Lab. Here is my understanding so far:

Method1: Signal Generator + Power Meter Generate about 0 dBm of power from the Signal Generator and measure the same on the Power Meter e.g. assuming 10 dB pads + cable loss 5 dB = -15 dBm (this the total cable path loss).

Method2: Network Analyzer Connect the 2 ends of the paths to Port 1 & Port 2 of the Network Analyzer and measure the S21 parameter.

I kind of get the idea of Method 1 but do not understand the Method 2 much.

\$\endgroup\$
2
\$\begingroup\$

When you're using a network analyser to measure S21, it generates a signal on port 1, and receives a signal on port 2. S21 is a measurement of the ratio of those signals.

While a network analyser does much more than a signal source and power meter can (S21 is the complex ratio), for this cable loss application, if you display simply the magnitude of S21, you can regard it as a signal source and power meter in a single box.

\$\endgroup\$
  • \$\begingroup\$ So it would mean, I can configure my Network Analyzer to a particular Frequency & Power level, which would be incident on the Port 2. And the Network Analyzer should provide an S21 measurement which is loss for that Cable Setup connected between Port 1 & 2. And this can simply be repeated for multiple frequencies. \$\endgroup\$ – eecs Mar 9 '17 at 7:18
  • 1
    \$\begingroup\$ Chaee, one of the jobs of a network analyzer is to do thing for many frequencies in succession. You might really want to read an intro to network analyzers. \$\endgroup\$ – Marcus Müller Mar 9 '17 at 14:34