The R&S HMC8015 power analyzer datasheet and technical data list:

Range configuration

            CF3      Peak
voltage     600 V    ±1800 V
current     100 mA   ±300 mA

Measurement accuracy (±reading in % ±peak range in %)

Frequency         ActivePower   
45Hz < f < 66Hz   0.05 + 0.05

voltage, current: F = frequency in kHz

Additional errors

Power factor < 1 | ±(0.2 + 0.2 *F)%, only for active power


Does this mean that when range is set to 600 V, 100 mA and measuring these values:

  • Urms = 229.96 V
  • f = 49.986 Hz
  • Irms = 15.33 mA (with peaks to 230mA)
  • P = 0.064 W
  • Q = 3.525 var
  • Lambda = 0.018

That the measurement accuracy for active power can be calculated as follows:

  1. 0.05% for reading + 0.2% for power factor < 1 = 0.25% of reading

plus

  1. 0.05% for peak range + 0.2*F for power factor < 1 = 0.05 + 0.2 * 0.049986 ≅ 0.06% of peak value for range

resulting in a nett measurement accuracy of:

  1. 0.06% of 0.064 W = 0.0000384 W

plus

  1. 1800 V x 0.3 A = 600 * 0.06% = 0.36 W

totalling to ≅ ±0.36 W of measurement accuracy?

\$\endgroup\$

1 Answer 1

0
\$\begingroup\$

Rohde & Schwarz Vertriebs GmbH does confirm that the accuracy calculation in the question above — of plus or minus 360 milliwatts for range 600V, 100mA at 50Hz — is correct for power factors below 1.