So I have this simple force measure circuit which consists of a strain gauge SG1 and the amplifier with the Wheatstone bridge, the variable resistor RV1 is set to 200Ohms.
If the strain gauge is fully deflected the circuit gives an output of 50mV.
How would I go about to calculate the imbalance voltage at the bridge.
What I did so far is I found the voltage that goes into the op-amp's pin3: $$R_{FSR}=\dfrac{1}{\dfrac{1}{1320}+\dfrac{1}{1320}}=660Ohms$$ $$V=\dfrac{680}{680+660}*9=4.57V$$ Any help will be appreciated.
-
1\$\begingroup\$ I am having a hard time understanding what you are asking here and an even harder time figuring out why you have the op-amp wired that way. It should be like this bing.com/images/… \$\endgroup\$– Trevor_GNov 30, 2017 at 1:23
-
\$\begingroup\$ there is nothing wrong with the circuit. \$\endgroup\$– super95Nov 30, 2017 at 1:29
-
\$\begingroup\$ Sure.. its fine till you hook any significant load on the output.. \$\endgroup\$– Trevor_GNov 30, 2017 at 1:47
-
3\$\begingroup\$ The Supply sensitivity is too high and you need a differential Amp on the output, when you should have it here. Change all 1k2 to 120 and add an LDO for 15mA into 120 Ohms or 1.8V then use a rail to rail Op Amp with >10K load using Trevor's link. You should get 500mV then full scale then and less CM error and gain error. \$\endgroup\$– Tony Stewart EE75Nov 30, 2017 at 1:53
-
2\$\begingroup\$ Your gain is ~200 so the full scale input is 250uV making it very prone to noise and errors. So there are a few things wrong \$\endgroup\$– Tony Stewart EE75Nov 30, 2017 at 1:59