My DMM has a accuracy of 0.0030% of reading + 0.0030% of range, when the voltage range is set to 100mV. If a measurement with this DMM gives 0.1mV, then is the true input voltage between 0.1mV - 3.003uV and 0.1mV + 3.003uV? Or is there a different factor than the uncertainty that I have to account for?
2 Answers
Your calculation is correct but the result is not really the uncertainty of your measurement, it's only the maximum instrument error.
If you want to calculate the uncertainty following the ISO standard you need to go a bit further. In this document you have a very good example:
Fluke DMM Uncertainty Calculation
If you follow the example presented with your data you'll get to the uncertainty values you're looking for. Give it a go and post your results, then we can let you know if your answer is correct.
If accuracy is ±0.0030% of a 100mV range, ±0.0030% * 100 mV = ±0.00003 * 100 mV= ± 3 μV
So if your example was 0.1 mV = 100 μV ± 3 μV
- Accuracy can be specified differently as a % of reading with a number of counts.
- It can be defined in terms of each part of error; gain, offset, non-linearity
normally examples like yours are expressed in parts per million, ppm
Features 34460A 34461A 34465Anew 34470Anew Digits of resolution 6½ 6½ 6½ 7½ 1 year DCV accuracy 75 ppm 35 ppm 30 ppm 16 ppm
low cost DMM's might be specified in terms of ±% of reading and ±counts
-
\$\begingroup\$ Could you please elaborate the 1e3 term in "±0.003/100 *1e3 = ± 0.03 μV"? I guess it's 1mV because the left hand side is dimensionless now \$\endgroup\$– HuismanCommented May 27, 2019 at 6:21
-
\$\begingroup\$ unit conversion %mV to μV. software calculators accept shorthand 1e3 for powers of 10 \$\endgroup\$– D.A.S.Commented May 27, 2019 at 6:24
-
1\$\begingroup\$ I think 1e3 should be 1e-3? And next, please elaborate how "a 100mV range" equals to "1e-3". It should at least have dimensions. \$\endgroup\$– HuismanCommented May 27, 2019 at 6:33