I'm studying semiconductor parameter analyzers. Those instruments are based on SMU (source and measuring unit) which can work either in the V-mode (they force a voltage on the device under test and measure the corresponding current) or in the I-mode (they force a current in the DUT and measure the corresponding voltage). Suddenly, it appears this slide:
It is just one slide and no further information is provided, but I cannot understand how this circuit works. On the web I found different configurations of this circuit. I might surmise that the "force" and "sense" words in the picture make reference to the "source and measurement" properties of the SMU. Moreover, why do we have a variable battery? Why do we have two "force" wires (if I force a current through Rf1 like in the picture, I don't need another "forcing" action on the DUT through Rf2)? Why do we have two "sense" wires?
Thank you
    \$\begingroup\$ This is a confusingly drawn circuit. I don't know how the actual internals of a four-wire ohmmeter work, but this is certainly far from how I'd design it with my understanding (which I think is fairly good) of the principles. \$\endgroup\$ – Hearth Jun 15 '19 at 18:14
    \$\begingroup\$ I agree with @Hearth, this is a confusing diagram but the concept is essentially the same as a 4-wire ohmmeter measurement. You should be able to find lots of better resources online. \$\endgroup\$ – Elliot Alderson Jun 15 '19 at 19:10
  • \$\begingroup\$ While the question seems valid on conceptual level, merely googling "Kelvin type connection" gives Wikipedia article covering this topic exactly. en.wikipedia.org/wiki/Four-terminal_sensing \$\endgroup\$ – Agent_L Jun 16 '19 at 7:37
9
The diagram is really confusing. In the 4-wire connection below on the right, for resistance measurement, the "force" wires are the ones (with red arrows) providing the current for the resistor under test. Through the "sense" wires only the current required by the voltmeter is circulating, causing a much lower drop.
In this example both current and voltage are being measured but the voltage source and current measurement could be replaced by a known current source. With the circuit on the left when you calculate the resistance from the current and voltage values you are actually calculating the resistance of the probe wires added to the DUT.
Regarding the "force voltage" "measure current" you mention, imagine if on the circuit on the left you didn't have the voltmeter. You would measure the current passing through the DUT, but the known voltage would be applied to the series association of the DUT, probe wires and the current meter. This would make the error even worse.
| improve this answer | |
    \$\begingroup\$ Those diagrams are superior. \$\endgroup\$ – DKNguyen Jun 15 '19 at 20:34
Correct. Each SMU can output a "force" signal (the excitation signal), and it can measure ("sense") the effects of that signal upon the device under test (DUT).
The variable battery is called a "ground unit" (GNDU). It is an active circuit that produces a very precise reference potential for the parametric measurement. The circuit ground is usually too noisy to be useful for ultra-precise voltage and current measurements. For example, a Keysight Technologies B1505A Power Device Analyzer can measure currents down to "sub picoamp" levels (<1E-12 amps). This level of measurement resolution would be impossible without an ultra-clean, actively-driven "virtual ground" reference potential.
Consider the circuit shown in Figure 1. Voltmeter VM1 has very high input impedance; therefore, all (or nearly all) of I1's current flows through the two test leads and the DUT.
Figure 1. 2-wire resistance measurement.
The voltage measured by voltmeter VM1 is
$$ VM1 = I1\,(R_{TestLead} + R_{DUT} + R_{TestLead}) $$
Current source \$I1\$ outputs a known (calibrated) current level. Voltmeter VM1 measures the voltage across the two test leads and the DUT and then the device analyzer uses Ohm's law to calculate the measured value of the DUT's resistance \$R_{DUT(meas)}\$:
$$ \begin{align*} R_{DUT(meas)} &= \frac {VM1}{I1}\\ &= \frac {I1\,(R_{TestLead} + R_{DUT} + R_{TestLead})}{I1}\\ &= R_{TestLead} + R_{DUT} + R_{TestLead} \end{align*} $$
EXAMPLE 1
The resistance in each test lead is \$100\,m\Omega\$, and the DUT's true resistance value is \$R_{DUT}=1\,k\Omega\$. In this case test the resistance in the two test leads introduces an error of about 0.02% to the measurement of the DUT's value.
$$ \begin{align*} Error\% &= \frac {Measured-True}{True} \times 100\\ &= \frac {R_{DUT(meas)}-R_{DUT}}{R_{DUT}} \times 100\\ &= \frac {(0.1+1000+0.1)\,\Omega - 1000\,\Omega}{1000\,\Omega} \times 100\\ &= 0.02\,\% \end{align*} $$
Therefore, if the resistance in the test leads is much less than the DUT's resistance, then we can ignore the test lead resistance terms and calculate \$R_{DUT}\$ as
$$ R_{DUT(meas)} \approx \frac {VM1}{I1} \bigg\rvert_{R_{TestLead} \lll R_{DUT}} $$
If, however, the DUT's resistance is very small—e.g., a few ohms or less, then the test lead resistance cannot be ignored because it adds significant error into the measurement of the DUT's resistance.
EXAMPLE 2
The resistance in each test lead is \$100\,m\Omega\$, and the DUT's true resistance value is \$R_{DUT}=1\,\Omega\$. In this test case the resistance in the two test leads introduces an error of about 16.7% to the measurement of the DUT's value. And this error doesn't include the voltmeter's own measurement error, which further increases the overall uncertainty in the measurement.
$$ \begin{align*} Error\% &= \frac {Measured-True}{True} \times 100\\ &= \frac {R_{DUT(meas)}-R_{DUT}}{R_{DUT}} \times 100\\ &= \frac {(0.1+1.0+0.1)\,\Omega - 1.0\,\Omega}{1.0\,\Omega} \times 100\\ &= 16.7\,\% \end{align*} $$
To improve the device analyzer's measurement accuracy when measuring small resistances, a "4‑wire" Kelvin measurement (Figure 2) must be used instead of the "2-wire" connection shown in Figure 1.
Figure 2. Kelvin 4-wire resistance measurement.
Current source \$I1\$ outputs a known (calibrated) current level. Recall that voltmeter VM1 has very high input impedance, and therefore almost no current flows through VM1. Likewise, almost no current flows through the "sense" test leads (R_SENSE), and therefore there is no voltage change (voltage drop) across the sense test lead resistance R_SENSE, \$V_{R_{SENSE}}\approx0\,V\$, which means the voltmeter is measuring the voltage at the DUT's input terminals:
$$ VM1 = \frac {I1\,R_{DUT}\,R_{VM1}}{R_{DUT}+2R_{SENSE}+R_{VM1}} $$
Note that if the voltmeter's input impedance \$R_{VM1}\$ is very high, then via L'Hôpital's rule:
$$ \lim_{R_{VM1}\rightarrow \infty }VM1 = I1\,R_{DUT} = V_{DUT} $$
The measured voltage in a 4‑wire Kelvin measurement, when used to calculate the DUT's resistance, yields a calculated resistance value that is much closer to the DUT's true resistance value when compared to a 2‑wire measurement. If the device analyzer calculates the DUT's resistance value using
$$ R_{DUT(calc)} := \frac {VM1}{I1} = \frac {R_{DUT}\,R_{VM1}}{R_{DUT}+2R_{SENSE}+R_{VM1}} $$
then the nominal percent error in the Kelvin measurement of the DUT's resistance due to circuit loading by the two sense test leads and the voltmeter's input impedance is
$$ \begin{align*} Error\% &= \frac {R_{DUT(calc)}-R_{DUT}}{R_{DUT}} \times 100\\ &= - \frac {R_{DUT} + 2 R_{SENSE}} {R_{DUT} + 2 R_{SENSE} + R_{VM1}} \times 100 \end{align*} $$
EXAMPLE 3
Given the resistance in each sense test lead is \$R_{SENSE}=0.1\,\Omega\$, the DUT's true resistance is \$R_{DUT}=1\,\Omega\$, and the voltmeter's input impedance is \$R_{VM1}=1\,G\Omega\$. The series impedance of the two sense test leads and the voltmeter, when placed in parallel with the DUT, introduces a loading error of -0.00000012% into the calculated value for the DUT's resistance.