The continuity shows us the shorts, but many times I guess, even if the wire has a resistance of up to 50 ohm, the continuity will beep to indicate a short.
Is it always 50 ohm, or what parameter should we look for in the manual?
The continuity shows us the shorts, but many times I guess, even if the wire has a resistance of up to 50 ohm, the continuity will beep to indicate a short.
Is it always 50 ohm, or what parameter should we look for in the manual?
It is not a standard value from meter to meter. Different models will beep at different resistances.
An example in the comments were the fluke models 175,177 and 179. In the datasheet for these, you can see on page 12 that the meter beeps at <25 ohms.
The multimeter I use is a Tenma 72-7732A, and in the datasheet for this model, on page 35 it is stated that the beeper will sound for conditions less than 50 ohms.
In the popular EEVBlog multimeter, page 25 of the Datasheet states the continuity threshold is between 30 and 480 ohms.
These few examples are enough to determine that there is no standard value. If the information cannot be found in the datasheet of your particular model, then getting a few resistors between 10 - 250 ohms and measuring them should be enough to tell you the threshold.
An easier way, as pointed out by @HarrySvensson in the comments is to turn a potentiometer/rheostat until the beeping stops/starts and measure the resistance.
I'm not aware of any standard resistance values for the continuity mode of a multi-meter.
However, I would personally say, there is an unofficial industry expectation that the continuity beeper should not trigger on a pn-junction.
Assuming the meter is using an excitation current of 1 mA and choosing a conservative IV curve for a pn-junction (assume Id < 100 uA at say 200 mV - 300 mV). By Ohms law, that would give you an upper continuity threshold of 200 \$\Omega\$ to 300 \$\Omega\$.
The continuity resistance threshold is not standard.
The exact value is certainly quoted in the datasheet/manual, if the instrument comes from a reputable manufacturer. If you have a cheap, unknown instrument then you can try a few resistors, and check which one beep, and which one do not beep.
Some expensive multimeters allow you to manually set the threshold, on my Keithley 2000 the threshold can be set anywhere between \$1\ \Omega\$ and \$1\ \text{k}\Omega\$.
I was curious as well. Just tested it on my Fluke 179 DMM with a relatively fine adjustable 500 Ω multiturn (8-turn) potentiometer.
simulate this circuit – Schematic created using CircuitLab
Data sheet says “Product beeps at <25 Ω, beeper turns off at >250 Ω; detects opens or shorts of 250 μs or longer.” See page 20 - Fluke 179 data sheet
My copy has the following behavior (quite different than specified in the data sheet):
It beeps as at <47 Ω, beeper turns off at >82 Ω
As a side note: I also tested the behavior of the pn-junctions on a BC547 NPN and BC557 PNP transistor with a continuity test. For all different junctions (forward direction) there was no beep, but a resistance of about 700 Ω is displayed on the LCD.