9
\$\begingroup\$

I am not well versed with transmission line theory so if you can redirect me to relevant material I'd be grateful. So I used Agilent 4294A to find resistance of a 2 metre long shielded twisted pair cable (BELDEN 3105A E34972 1PR22 SHIELDED) and the resistance across frequency looked something like

Resistance across frequencies

with a discontinuity at 5MHz. At 4.99 MHz it was about 2.04 Ohms and 23.5 Ohms at 5.01 MHz. This trend was there in impedance as well. I feel I'm missing something fundamental here.

\$\endgroup\$
25
\$\begingroup\$

Your tooling seems to be the cause there, not the cable. From https://www.keysight.com/main/editorial.jspx?cc=US&lc=eng&ckey=1428419&nid=-32775.536879654&id=1428419

The 4294A extends its measurement frequency range up to 110 MHz by terminating each measurement terminal with 50 ohm in order to eliminate the resonance of test leads (including leads inside the 4294A). The measurement discontinuity is caused by the change in termination impedance at 15 MHz when the ADAPTER is set to NONE or at 5 MHz when it is set to 1m or 2m. The measurement discontinuity can be removed by performing LOAD compensation.

New contributor
William Lieurance is a new contributor to this site. Take care in asking for clarification, commenting, and answering. Check out our Code of Conduct.
\$\endgroup\$
  • 1
    \$\begingroup\$ +1 for actaully looking in the manual, rather than guessing \$\endgroup\$ – Neil_UK 2 days ago
20
\$\begingroup\$

Something as simple as a cable does not have discontinuities like that.

There may be a clue in the fact the problem occurs at a nice round number, 5MHz. Is this a place where your test set changes ranges? Maybe it changes output amplifier, or filter, and one of them is broken or damaged.

The fact that you've quoted measurements at 4.99MHz and 5.01MHz without listing them hints that you have more data hidden that might throw light on what's going on. Listing spot measurements at a few selected frequencies is fine when everything is behaving itself, but not when you're hunting for an anomaly. The detail of the response adjacent to 5MHz will be very valuable.

Please edit your question with a plot of all the data you have taken, which may allow us to make better guesses. A connection schematic to show exactly how the cable is connected to the analyser would be useful as well.

\$\endgroup\$
  • 5
    \$\begingroup\$ Sure. I'll try to get a plot of the response. \$\endgroup\$ – Vibhore Jain Aug 14 at 8:44
1
\$\begingroup\$

Consider the cable (I assume coax) as a string of small inductors with capacitors at the junction of each pair of inductors to ground (the shield). At low frequencies the inductors act as they would with near DC signals (a wire) and the capacitors would be near opens at the near DC signals.

As the frequency goes up the inductors have more reactance and the capacitors have lower impedance, eventually forming effectively a series of LC filter poles. At some frequency the combined filter characteristics will become pronounced, especially with an unterminated (50-75 Ohms) line. Add the correct termination resistance and things should look a lot better behaved. Most coax cables do have an upper limit of usefulness due to the inter electrode capacitance.

New contributor
PAT is a new contributor to this site. Take care in asking for clarification, commenting, and answering. Check out our Code of Conduct.
\$\endgroup\$
  • 5
    \$\begingroup\$ 5MHz isn't particularly fast for twisted pair cable. So I don't think transmission line effects will cause such a large discontinuity. \$\endgroup\$ – Navin 2 days ago

Your Answer