I both check via voltmeters and osc. but oscilloscope is showing the values 10 times bigger. For example 5V pulse should be 2.5 mean and voltmeters shows exactly the same but osc. shows 25 V. What can be the reason?
2 Answers
Sounds like the scope is expecting a 10x probe (which attenuates the signal 10x), but you are using a 1x probe instead, therefore the scope is multiplying the signal by 10 when it shouldn't be. Check your settings.
Look for the "Probe" settings for your o-scope. There should be an option to set the probe to a "10x". Seriously. It's that straightforward.
-
-
3\$\begingroup\$ @DavidKessner I think you have it backwards -- I think the scope is already set to 10x and he is using a 1x probe, so he either needs to set the scope's 1x/10 setting to 1x, or switch to a 10x probe. \$\endgroup\$– tcrosleyDec 19, 2012 at 2:20