Does anyone knows why an over compensated probe gives voltage spikes bigger than the maximum voltage of the input signal (or bigger Vpeaks?)

These spikes looks to me like a parasitic inductance but I read that the only reason that the variable capacitor is placed on a probe is to compensate the 9Mohm resistor which creates a low pass filter with the input capacitance of the oscilloscope.

These spikes look like a voltage that comes from a resonance frequency between the capacitors and maybe the parasitic inductance which is formed through the probe's cable.

If the probe is under compensated then we have low pass filter which cuts the high frequencies so why there is a ramp shape at the transitions of the signal?

enter image description here

\$\endgroup\$
1
  • 2
    \$\begingroup\$ It doesn't. It's a x10 probe, so the spike is greater than 1/10 the input. \$\endgroup\$ Mar 15, 2021 at 19:00

2 Answers 2

4
\$\begingroup\$

It is a resistive voltage divider in parallel with a capacitive voltage divider:

enter image description here

The output is compared with 1/10 of the input (so, it is not above the input):

enter image description here

The smaller capacitor at the probe is adjusted to match the voltage ratio determined by the resistance of the 10x probe (9M) and the oscilloscope input (1M).

EDIT: to address your comment on frequency response

As for the frequency response, considering only these characteristics and not the others that go beyond your question, you can consider the 10x probe as a voltage divider that deviates from its DC ratio as frequency increases, if not compensated properly.

enter image description here

\$\endgroup\$
2
  • 1
    \$\begingroup\$ So if I understand right we could say that when the probe is undercompensated the impedance of 9Mohm parallel with Capacitor (R1-C1) at high frequencies is much bigger and keeps a bigger part of the voltage on it from the oscilloscope input (1MΩ - R2-C2), so the ratio 1/10 is not exist. On the other way when the probe is overcompensated the impedance of 9Mohm parallel with Capacitor (R1-C1) at high frequencies is much smaller so the ratio 1/10 is not exist and oscilloscope receives bigger voltage than 1/10 ratio and for that we see bigger spikes. \$\endgroup\$
    – 2 X
    Mar 15, 2021 at 20:50
  • 1
    \$\begingroup\$ Yes, you can think that these parallel associations degrade the "perfect ratio" of the voltage divider as the frequency increases. Of course there are other issues to consider, but, to address your question directly, that is a simple model that describes it. \$\endgroup\$
    – devnull
    Mar 15, 2021 at 21:17
0
\$\begingroup\$

The variable C on the probe is to compensate for the tolerance the dielectric shunt capacitance of the 10:1 probe.

The first diagram is amplified but shows the effect of excessive differentiation from high variable C in series to the 9C cable shunt capacitance. They intensionally use higher impedance probe coax and thus higher L/C or lower C for a given length to reduce the size of the tiny variable C/9. Thus the total probe load is the scope impedance x 10 with ~C/10 with both C’s in series to the load. You get C/10 by = (C/9 * C ) / ( C/9 +C) = C-probe for DSO input= C

Thus 1/(9+1) ratio of C’s is called a 10:1 capacitance transformer just like inductive turns ratio transformers is for low frequency, thus is just for high frequency balancing with the R ratios of the same in parallel with C’s. Except the probe is 9M and the DSO is 1M also with some capacitance. So the probe must be rebalanced if used on a different scope using to provided test port with a square wave.

Another variable is the ground lead inductance which will cause ringing near 20 MHz assuming ~0.5 to 1 nH were lead length resonating to probe coax capacitance .

  • This requires you to remove the ground clip and Probe clip and use a spring probe over the tip and barrel of the probe with 2 wire pins near signal + gnd in order to get the full BW of scope whether it is 50, 100, 200, or >1GHz

It gets a bit more complex than this as explained by TEK and HP on probe designs but essentially you get a high impedance matched 10M to 1MOhm broadband voltage divider.

\$\endgroup\$