I use a pocketVNA to measure the impedance of antennas and I also bought a couple of SMA connectors and a few cables to connect the antennas to the device.

I run a 2-point calibration of the device (according to the manual) by using the calibration kit. Please take a look at the screenshot for the calibration results.

enter image description here

I´m not sure if these calibration results are good or bad. What are the critical regions for such a calibration?

After the calibration, I have tried to measure the impedance of a know 868 MHz antenna, but the result doesn´t look good:

enter image description here

I would expect the peak at around 868 MHz and not 900. As I understand it the antenna should have a resistance of around 50 Ohms at 868 MHz. So why does the antenna has an impedance of around 10 Ohms?

Also, the result is getting really worse when the antenna is laying on my table instead of holding it into the air:

enter image description here

I don´t understand the results and I suggest that my handling of the device is wrong. How can I improve it? Are there some tips for using a VNA?

Update:

The calibration process is a simple 2-Port calibration with three different load types:

  • Open
  • Short
  • Load (Don´t know the resistance)

enter image description here

Please take a look at the photo to see my setup. I have done the whole calibration with the cable, because I want to use the cable during the measurements.

enter image description here

\$\endgroup\$
2
  • \$\begingroup\$ Is it actually 3-point calibration? Open, short and 50 ohm. Do you have pictures of your setup? \$\endgroup\$
    – Ralph
    Mar 13, 2022 at 19:19
  • \$\begingroup\$ @Ralph please take a look at my updated question \$\endgroup\$
    – Kampi
    Mar 13, 2022 at 20:51

1 Answer 1

2
\$\begingroup\$

calibration looks ok.

Maybe your table has some metal in it, so the antenna response changes when you move it?

Hope this helps, pVNA

\$\endgroup\$
3
  • \$\begingroup\$ Hi, thanks for your response. I have checked the manual and want to follow these steps. Which calibration kit should I use when I have ordered the original pocketVNA calibration kit (Custom,Agilent 7 mm, Agilent 3.5 mm) ? \$\endgroup\$
    – Kampi
    Mar 23, 2022 at 6:02
  • 1
    \$\begingroup\$ It is called "built-in". Bit unlucky name, I will rename it to "built-in to pVNA SMA calkit male" in the next version. \$\endgroup\$
    – pvna
    Mar 23, 2022 at 8:02
  • \$\begingroup\$ The problem was solved via email support. I have received the wrong calibration probe. My "open" probe was a "load" probe, marked as "open". I have calibrated the VNA again by leaving the port open instead of using the "open" probe and the results are looking way better. \$\endgroup\$
    – Kampi
    Mar 24, 2022 at 18:28