The derating curve is to avoid damage due to heating of the capacitors used in the input of the probe. This heating is a result of resistive and dielectric losses, and is proportional to the RMS voltage and frequency.
To stay within the power dissipation limit of the capacitors, the derating curve effectively plots the line of constant power, with a low frequency voltage limit driven by the breakdown voltage of the capacitors.
Specifically for this probe, if you are measuring an 800V RMS signal, the frequency must be below approximately 3MHz. Short term operation above this limit may be possible, but will be outside of the approved operation range, and "mileage may vary".
Measurements of sharp 1-off pulses is very unlikely to be an issue, despite containing high frequency elements, because average power is limited. It is the heating due to constant HF AC that is the key concern.