8
\$\begingroup\$

My understanding that when the terminals of an oscilloscope probe are shorted, the voltage level shown should be around zero.

Today, in my lab, I shorted both terminals and then connected the probe to a few of metal exposures on equipment laying around, whereby I saw some unnerving noise on the scope.

Pictures are attached below:

Probe terminals shorted Probe terminals shorted

Oscilloscope with probe terminals shorted Voltage reading on the scope

Probe terminals shorted AND connected to a metal surface Probe terminals shorted AND connected to a metal surface

Oscilloscope with probe terminals shorted AND connected to a metal surface. Noise is visible Voltage reading on the scope. Noise is visible

I have tried disconnecting the equipment from the mains, but the noise still remains visible on the scope.

What could be the reason?

\$\endgroup\$
1
  • \$\begingroup\$ Quite high noise levels. As @tobalt suggests, use the gnd-to-probe-tip arrangement of photo 1 to probe your environment - you should be able to identify noisy sources in and around your lab. \$\endgroup\$
    – glen_geek
    Commented Jun 23, 2023 at 12:52

3 Answers 3

11
\$\begingroup\$

This is due to imperfect common-mode rejection of the oscilloscope's input circuitry.

You initially short the probe tip and, there is a little bit of noise. This is to be expected because, all amplification systems are imperfect. Then, you connect the shorted probe tip to an alternative earth point. That alternative earth point might have several tens or hundreds of millivolts AC signal on it with respect to the oscilloscope's own ground and, you see a signal.

That is also to be expected.

If you used a differential probe (instead of an unbalanced scope probe) then it's much more likely that the waveform seen on the scope will be like your first scope image i.e. the probe shorted and not connected to anything.

Of course, the short makes a loop antenna and you can easily pick up mV to volts when testing such things as switch mode power supplies but, in you example, I believe that what you see is due to imperfect common-mode rejection of the scope's input circuitry.

\$\endgroup\$
18
\$\begingroup\$

Generally, the "short" you form with the clip forms a loop antenna which readily picks up magnetic interference when in proximity.

To confirm/exclude this: If you go close to the object without touching it, you would still detect the signal if it is due to magnetic pickup of the clip loop.

In contrast, if the signal appears only when in contact, a grounding current flows through the scope lead's outer contact and the finite impedance of this lead causes a detectable voltage drop that you are seeing. Even if the object you touch is itself earthed, it will still develop RF potentials, because of the ground loop formed between it and the scope.

\$\endgroup\$
0
6
\$\begingroup\$

Like Andy aka said, the issue is the lack of common mode rejection of the measurement setup. To some extent this is to be expected when not using a differential probe, but single-ended probes work well for many practical tasks.

If the input was shorted with 0 ohm impedance, the oscilloscope would see 0 volts input. But it isn't - all of the cables in the system have significant inductance:

schematic

simulate this circuit – Schematic created using CircuitLab

The cable inductance picks up noise and causes voltage drop. In the coaxial pair, the inner and outer conductor are closely coupled and noise voltage mostly cancels out. However, the few centimeters of the ground clip wire already presents a significant unbalanced inductance.

When the noise source is floating, there is not much voltage drop over this inductance. But when the noise is coupled through mains cabling, current will flow through the probe cable outer conductor. At the oscilloscope end, the outer terminal is connected directly to the power supply ground. Thus it presents a low impedance to any ground-referred noise source, and the noise voltage at oscilloscope ground terminal is small - most of the voltage is lost in the cable inductance.

That sounds like a good thing, but the positive terminal presents a high (typically 1 Mohm) impedance. The noise coupled there is not attenuated, and is detected by the input amplifier. Cancellation doesn't occur because the common ground causes attenuation at the negative terminal.

The source of the high frequency noise is probably some mains-powered device nearby. For me, it has usually been cheap LED lighting. Try disconnecting devices and see if affects the detected noise.

Switching to a shorter ground clip cable on the oscilloscope or to a ground spring will also give a big improvement. Here is a comparison I did with 10 MHz 20Vp-p square wave from a signal generator. White line is with 10 cm long ground clip cable, and yellow line is with ground spring (both shorted to the probe tip, like in original post):

Screenshot from oscilloscope screen, 400mVp-p noise on white trace and 20mVp-p on yellow.

\$\endgroup\$
5
  • \$\begingroup\$ @tobalt The system as total has poor common-mode rejection. At the oscilloscope input connector there is a true differential signal, but that is because of high-frequency impedance of the probe cable. \$\endgroup\$
    – jpa
    Commented Jun 24, 2023 at 6:26
  • \$\begingroup\$ The "outer shield" inductance should be labeled "ground clip", which is the source of the phenomenon. The two inductances as labeled have k = 1 (which should also be labeled; CircuitLab doesn't really have a way to couple inductors, sadly) so would give a good differential measurement. (The measurement can be improved somewhat by passing the probe cable through a ferrite bead, making the coupled inductance much larger than the ground lead inductance.) \$\endgroup\$ Commented Jun 24, 2023 at 15:38
  • \$\begingroup\$ @TimWilliams I agree that in >= 50 MHz measurements, the ground clip inductance is the most significant. I'm not convinced that it is true for this case, because the signal shown at 1µs time scale doesn't appear that high frequency - no way to be sure. I don't think there is much coupling the the probe cable as-is, but with a ferrite ring there could be and it should reduce the detected noise. \$\endgroup\$
    – jpa
    Commented Jun 24, 2023 at 16:55
  • \$\begingroup\$ Both signals are handily in the >10MHz range where ground clip is important, and the first could well be internal noise given the vertical scale, but could also be loop induction. Probe shield should be effective down to 10s of kHz, i.e. very well coupled indeed at these frequencies, unless it's well used and the ground is frayed (in which case the apparent noise will be intermittent as the probe is moved). \$\endgroup\$ Commented Jun 24, 2023 at 17:11
  • 1
    \$\begingroup\$ @TimWilliams Looks like you are correct - I did a test myself with ground clip vs. ground spring and the difference is huge (added image to post). I guess I'll be even more suspicious of ground clips from now on.. \$\endgroup\$
    – jpa
    Commented Jun 25, 2023 at 6:01