I'm a physics PhD student working on a project to use silicon photodiodes for detection of longwave infrared (6-10um) ultrashort pulses. (Specifically doing higher-order autocorrelation) Despite silicon having a bandgap of 1.12eV, it is possible to generate a nonlinear response via multi-photon and tunneling ionization in the silicon chip. Here is a simplified layout of the set-up:
The laser is first sent through a variable attenuator (details not relevant) and is then focused using a lens (again, details not relevant) onto the photodiode. The variable attenuator is used because the standard approach to characterizing the nonlinear response is to graph the output signal (voltage or photocurrent) as a function of input intensity (W/cm^2) like seen in the plot below:
I doubt that the details of the nonlinear signal generation will be germane to an answer to my question, but I thought it best to include it anyway.
The LWIR pulse going into the photodiode is extremely short (~50fs) so the detector should see it as being practically an impulse response.
In terms of electronics, my current set-up has a PIN photodiode (Thorlabs FDS100), a detector circuit (Thorlabs DET410), a 10ft 50-Ohm BNC cable, a BNC T-Plug (One arm goes to the PD and the other arm has a 50Ohm load), and finally an Agilent DSOX-2022A ocilloscope (1MOhm input port).
The problem I can't figure out is whether the signal I see on the O-scope is "representative" or not. Starting with very low intensity, I just start to see the signal peak out of the noise and it has a rising and falling edge a bit faster than that specified by the manufacturer. (~5ns actual, ~10ns spec). However, as I increase the intensity the rising edge stays the same, but the falling edges broadens significantly in a way that I would normally associate with a saturation effect. That being said, the voltage level is still only in the 100mV range when this broadening begins to occur and the Max Reverse Bias for the PD is 25V while the detector unit bias voltage is only 12V. Below is a drawing roughly showing this broadening effect:
If I increase the intensity and thereby the signal significantly I do begin to observe saturation effects on the oscilloscope (~3.6V) and this is accompanied by truly expansive broadening of the falling edge like we would normally anticipate.
So to put the question succinctly: Is there a reason, from an electronics perspective, that I should see this premature temporal broadening of the signal as I increase the intensity? Or is it possibly an actual physical effect?
For the sake of clarity, here are four images of the o-scope trace at different intensities:
1: Barely visible signal (5mV/div) 2: At the transition to this broadening effect (10mV/div) 3: Well into the broadening regime (100mV/div) 4: At saturation (1V/div)
Thanks for any and all responses!