Feb 21, 2018 at 16:12 | vote | accept | beard999 | ||
Jan 21, 2018 at 21:31 | comment | added | Dave Tweed | For example, on any of the ranges from 2mV/div through 50mV/div, you will be able to bring any voltage between +/-0.4V to the center of the screen using the vertical offset control. That's what "offset range" means. | |
Jan 21, 2018 at 19:22 | comment | added | beard999 | Can you be more explicit ? I don't really understand. The offset range bandwidth limit is the maximum voltage allowed at the input of the oscilloscope or the maximum input voltage is 300Vpeak ? | |
Jan 21, 2018 at 19:17 | comment | added | Dave Tweed | "Offset range bandwidth limit" is actually two separate specifications: "Offset Range" (the top line) tells you how far the trace can be offset from zero for the various vertical scale settings. "Bandwidth Limit" (the bottom line) simply tells you the -3dB bandwidth of the scope overall. | |
Jan 21, 2018 at 19:03 | comment | added | beard999 | So the "offset range bandwidth limit" doesn't mean the maximum input voltage ? | |
Jan 21, 2018 at 18:02 | comment | added | beard999 | Thanks for the reply. What means "offset range bandwidth limit" ? I saw this in the manual: imgur.com/LB1cvYr | |
Jan 21, 2018 at 17:13 | history | edited | Dave Tweed | CC BY-SA 3.0 |
added 53 characters in body
|
Jan 21, 2018 at 17:08 | history | answered | Dave Tweed | CC BY-SA 3.0 |