I'm reading about spectrum analyzers from Keysight's AN 150. On page 12 they explain how multiple mixing stages are used to obtain narrower IF filters. The block diagram offered is the following one:
They state that the tuning equation would then be:
$$f_{sig}=f_{LO1}-(f_{LO2}+f_{LO3}+f_{final \ IF})$$
And that
$$f_{LO2}+f_{LO3}+f_{final \ IF} = f_{first \ IF}$$
So they say:
Simplifying the tuning equation by using just the first IF leads us to the same answers.
My doubts
I tried to deduce the first equation, but I'm getting a sign wrong. My reasoning was the following. I have a sinusoid of frequency \$f_{sig}\$. After the first mixing stage, I get two components at \$f_{LO1} + f_{sig}\$ and \$f_{LO1} - f_{sig}\$. The first one will be filtered out, so after the second mixing stage we have \$f_{LO2} - f_{LO1} + f_{sig}\$ and \$f_{LO2} + f_{LO1} - f_{sig}\$. Again, the first one is filtered out and after the third and last mixing stage, and after the filtering too, we have that the following equality must be satisfied to generate a response on the display: $$f_{LO3}-f_{LO2} + f_{LO1} - f_{sig}=f_{final \ IF}$$ Reordering: $$f_{sig}=f_{LO1} -(f_{LO2}-f_{LO3}+ f_{final \ IF})$$ So in my equation, \$f_{LO3}\$ appears with a negative sign, but on the AN it doesn't. Where did I make the mistake?
The other question I have is regarding the second equation shown above: \$f_{LO2}+f_{LO3}+f_{final \ IF} = f_{first \ IF}\$. Is this a design requirement? I mean, must this equality be satisfied in order for the spectrum analyzer to work properly? Could the frequencies of each mixing stage be any combination so that the equality holds?