I want to know the current spectral density of a Keithley 6221 current source.
The datasheet above provides values in RMS for a given Bandwidth from \$10~\text{Hz} - \text{BW}\$, not clearly specifying the bandwidth \$\text {BW}\$. I am not exactly sure what Output Response Bandwidth (BW) into Short means, but figuring from the \$\text {BW}\$ in brackets I used it as bandwidth. In a first approximation the RMS value can be calculated from the spectral density \$\text {SD}\$ by $$X_{RMS}=SD\cdot\sqrt{\text{BW}}\cdot\sqrt{\text{ENBW}}~~~~~~~~~~~(1)$$ with the equivalent noise bandwidth \$\text {ENBW}\$ (rectangular approximation valid for high bandwidths). Assuming the above values for the 100 mA range this yields $$\frac{20e^{-6}\text{A}_{\text{RMS}}}{\sqrt{1e^6~\text{Hz}}\sqrt{\text{ENBW}}}=20\frac{\text{nA}}{\sqrt{\text{Hz}}},~~~~~~~~~~~(2)$$ which seems a lot to me. Note that a \$\text{ENBW}\$ anywhere from 1 (brick wall filter) to 1.57 (1 Pole) does not change the result significantly and can thus be neglected.
Is this the correct way to determine the spectral density?
I read through the thread Determining noise spectral density, which states that the spectral density cannot be calculated due to lack of information. If the bandwidth I used is correct, this is not the case here.
Sideinformation: I feed a 3 Ohm inductance with the current source and I can measure the voltage density over the inductance with a spectrum analyzer. I would like to get the calculation and the measurement into accordance. The voltage noise can be calculated from the current noise and the resistance, once the current noise is known.