I have been working in a calibration system for a multimeter I'm developing.

To do this I'm calibrating the reading values using Fluke meters (basically my instrument has to match the Fluke ones).

However at college I was told that this is not the right way to do this, that I'm assuming the Fluke meters are on the right and have no error margin, so I was told to do the calibration using normalized loads.

I don't know what are or where to find these "normalized loads" so I came here to ask, what's the right way to calibrate a measurement instrument? Where can I find these kind of references?

\$\endgroup\$
3
  • 1
    \$\begingroup\$ the correct way to calibrate is to use a standard, standard voltage source, standard resistor etc, or more commonly to pay someone else to do it, as standards can be expensive. \$\endgroup\$ Commented Jan 5, 2016 at 22:47
  • \$\begingroup\$ You should feel more embarrassed to ask here than waiting for your prof to come back and get a decent explanation that you understand. Ask the prof or are you too scared. \$\endgroup\$
    – Andy aka
    Commented Jan 5, 2016 at 22:55
  • \$\begingroup\$ @Andyaka I will see him again in about 3 weeks so i need to start looking for any source so i can do this test meanwhile, i edited the content of the post either way \$\endgroup\$
    – GoatZero
    Commented Jan 5, 2016 at 23:11

1 Answer 1

2
\$\begingroup\$

There are two entirely separate issues here, precision and accuracy.

If you adjust one meter to read the same as the other, you are adjusting the second to have the same precision as the first. That is, if A reads 9.781 volts and B reads 9.781 volts, each is providing the same precision.

However, if both are measuring a 5.221 volt source, both are equally inaccurate.

So - are you trying to calibrate your meter to "the fluke ones" (matching precision) or are you trying for accuracy? If the first, you can use almost any standards you like as long as you are sure that they don't drift with time, temperature, or anything else. You measure the standards with "the fluke ones", record the readings, then adjust your other meters to match.

If you're trying for accuracy, you'll need to do one of two things: either buy or build a good reference, or rent or borrow a very good, high-resolution meter to find the errors in "the fluke ones". You should be able to find either solution at a test equipment house. There are many online.

\$\endgroup\$
1
  • 1
    \$\begingroup\$ That's not really a good description of precision. Precision means if the voltage changes by a small amount (say, 0.001 V) does the measurement instrument detect the change. If the reference meter measures 9.781 and the meter being tested measures 9.781, it doesn't mean the precision is 0.001 unless a 0.001 V change in the actual voltage causes the readings to change. \$\endgroup\$
    – The Photon
    Commented Jan 6, 2016 at 3:42

Your Answer

By clicking “Post Your Answer”, you agree to our terms of service and acknowledge you have read our privacy policy.

Not the answer you're looking for? Browse other questions tagged or ask your own question.