I'm searching for some simple 3-wire RTD sensor to be measured using an ADC. I found some options, but I'm kinda confused with how does they works and which one is the best.

First, there is the Weasthone bridge configuration. I found two variations of this circuits, and I cant understant how this compensates the wire resistances Rw.

schematic

simulate this circuit – Schematic created using CircuitLab

schematic

simulate this circuit

This second one, I think I understand how does it works, but I dont know how to implement this. I can subtract VM1 - 2*VM2 using an AmpOP, but how do I implement a constant current source?

schematic

simulate this circuit

\$\endgroup\$
1
  • \$\begingroup\$ Analog Devices have some app notes. \$\endgroup\$
    – Kartman
    Commented Apr 19, 2021 at 13:34

1 Answer 1

1
\$\begingroup\$

Okay, number 1- think of the resistors to the much higher excitation voltage as constant current sources:

enter image description here

The two currents are chosen to be the same (nominally) so the voltages across Rw cancel out with a commmon-mode voltage of 2IRw.

I don't think the middle one works.

The last one, yes you can subtract double the voltage with a differential amplifier.

It's also possible to do something like this:

schematic

simulate this circuit – Schematic created using CircuitLab

Where R5 is chosen to pass double the current of R6.

There are many ways to skin this particular cat. You would also find various application notes from IC manufacturers that emphasize the use of expensive boutique chips they happen to supply.

Since the resistors are not perfect constant current sources, the cancellation will not be perfect and it will vary with the measured temperature but that's okay for most industrial situations as the wire resistance is usually much lower than the sensor resistance anyway and it's not perfectly matched since they are physically different wires of similar length. It's not the mathematically-elegant 4-wire situation, but more than good enough and used by 99.9% of all industrial installations.

\$\endgroup\$